Search found 1152 matches
- Wed Apr 02, 2025 1:39 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Electrode labels when checking offsets in ActiView
- Replies: 1
- Views: 81
Re: Electrode labels when checking offsets in ActiView
Use the CFG file for 64 channels with 1020 labeling, best regards, Coen (BioSemi)
- Tue Feb 11, 2025 10:21 pm
- Forum: BioSemi Hardware
- Topic: Updating firmware for high speed
- Replies: 2
- Views: 55576
Re: Updating firmware for high speed
Contact info@biosemi.com
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
- Tue Feb 04, 2025 3:52 pm
- Forum: BioSemi Hardware
- Topic: Need Advise!!
- Replies: 1
- Views: 10526
Re: Need Advise!!
Since the skin is not abraded, disinfection is generally not needed. When disinfection is preferred (as was the case during the Covid-19 outbreak), then follow the instructions as shown in https://www.youtube.com/watch?v=-CZx3_tbrYM . Do not use hydrogen peroxide since it may damage the headcap fabr...
- Tue Feb 04, 2025 3:35 pm
- Forum: BioSemi Hardware
- Topic: Compatibility with Transcutaneous vagus nerve stimulation (tVNS)
- Replies: 1
- Views: 11200
- Tue Dec 17, 2024 7:50 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Access Offsets from.bdf files
- Replies: 1
- Views: 42047
- Fri Nov 15, 2024 7:45 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 57147
Re: Error 5000
Contact your local BioSemi representative
- Fri Nov 15, 2024 6:36 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 57147
Re: Error 5000
ADC6 is an Mk1 AD-Box and does not work with the red receiver. Your white receiver does work with the USB Trigger Interface (the trigger ports of white and red receivers are identical).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Nov 15, 2024 1:44 pm
- Forum: BioSemi all other topics
- Topic: Real-time processing of EEG signals from Active3
- Replies: 2
- Views: 15054
Re: Real-time processing of EEG signals from Active3
ActiView for ActiveThree can stream the data from the ringbuffer via a TCP/IP connection. You can consider to develop a MatLab client to read this real-time data stream.
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
- Tue Nov 12, 2024 5:59 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 17944
Re: ActiView10 not working with USB2 receiver with Firmware4.0
It is normal operation that only trigger lines 15 (start/stop) and 16 (1 sec pulses) are transmitted from Master to Slave. The stimulation triggers should indeed be connected to the Master. The trigger lines 1-14 of the Slave are ignored when the Master and Slave files are merged into a single hyper...
- Mon Nov 11, 2024 3:01 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 17944
Re: ActiView10 not working with USB2 receiver with Firmware4.0
ActiveTwo (Mk2 and Mk2.5) only works with ActiView 9.xx ActiveThree only works with ActiView 10.x Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling: Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate Slave: ActiveTwo with ActiView 9.03 on 2 kHz sam...
- Wed Nov 06, 2024 6:30 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Errors 7000 and 7001
- Replies: 2
- Views: 13337
Re: Errors 7000 and 7001
The above CFG is for ActiveTwo.
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
- Thu Oct 31, 2024 3:41 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 16810
Re: hardware filters
Which "analog box" exactly are you talking about ?
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
- Thu Oct 31, 2024 3:25 pm
- Forum: BioSemi all other topics
- Topic: BIOSEMI Trigger and EEGLAB reading events
- Replies: 1
- Views: 11694
Re: BIOSEMI Trigger and EEGLAB reading events
ActiveTwo Mk1 and Mk2 (as well as the new ActiveThree) systems use exactly the same BDF file format (https://www.biosemi.com/faq/file_format.htm). In particular, storing of the 16 trigger lines in the Status channel is the same for all systems. The only difference between Mk1 and Mk2 files is the MS...
- Thu Oct 24, 2024 6:38 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Older Versions Repository? Looking for ActiView Version 705
- Replies: 1
- Views: 11813
Re: Older Versions Repository? Looking for ActiView Version 705
Use the current 9.02 version, it works with every ActiveTwo system (even as old as from 2002). You might have to upgrade the firmware of your USB-receiver to use it with Windows 10/11, see https://www.biosemi.com/faq/install_USB.htm.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Oct 02, 2024 1:52 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 16810
Re: hardware filters
The BioSemi system has DC amplifiers, there is no high-pass filter. Check the signal in ActiView with the (display) filters switched off. You have either a high-pass in your file reader, or you have capacitively coupled the trigger line to the Analog Input Box.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Sep 25, 2024 1:32 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 24000
Re: Hyperscanning with LabStreamingLayer
With the triggered hyperscanning method, the data files from the two (or more) systems are merged into a single hyperscanning file offline. So it is not possible to stream hyperscanning data during the recording.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Sep 24, 2024 3:19 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 24000
Re: Hyperscanning with LabStreamingLayer
According to the LSL website it should work with ActiveTwo (2 and 2.5), but that's all I know.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Mon Sep 23, 2024 2:30 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 24000
Re: Hyperscanning with LabStreamingLayer
You can Hyperscan with the Active3 and Active2.5 system using the method described on https://www.biosemi.com/faq/Hyperscanni ... erface.htm .
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
- Mon Sep 09, 2024 6:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 57147
Re: Error 5000
Send the red receiver to BioSemi, we will replace it with a white one (we still have a few).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 5:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 57147
Re: Error 5000
ADC11 are Mk1 systems (20 years old), they do not work with the new red receiver. You should use the older type white receiver.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 4:15 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 57147
Re: Error 5000
What is the serial number of your AD-box ?
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Aug 06, 2024 2:33 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: How to open .cgf
- Replies: 1
- Views: 14552
Re: How to open .cgf
The CFG file stores all ActiView control settings, the channel labels, and a few system settings (e.g. how response switch signals are linked to triggers). ActiView opens with the default.cfg file. On the About/Configure tabpage, you can read a CFG file (for example a CFG with 32-chan settings and 1...
- Tue Aug 06, 2024 2:03 pm
- Forum: BioSemi Hardware
- Topic: CMS/DRL are unstable
- Replies: 1
- Views: 19293
- Mon Jul 15, 2024 6:03 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Detailed review on the concept of offset?
- Replies: 4
- Views: 22023
Re: Detailed review on the concept of offset?
Google on "opamp buffer output impedance". With active electrodes, the electrode impedances in the interference paper (Ze in Fig. 1) are reduced from tens of kOhm to less than an Ohm, and the magnitude of interference is reduced with the same ratio.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Jul 12, 2024 7:15 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Detailed review on the concept of offset?
- Replies: 4
- Views: 22023
Re: Detailed review on the concept of offset?
1. What is the definition of offset? The electrode-gel-skin interface forms a galvanic half-cell. The half-cell potential is determined by the electrode metal. The offset voltage is the potential difference between two of these half-cells. When the electrodes are identical (same metal and perfectly ...
- Fri Jun 07, 2024 2:36 pm
- Forum: BioSemi all other topics
- Topic: Modify ActiView Configuration
- Replies: 1
- Views: 20766
Re: Modify ActiView Configuration
The above is not a standard option in ActiView that can be enabled with a CFG setting. Bipolar display for all channels would require changing the (open source LabVIEW) source code.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Thu May 30, 2024 1:38 pm
- Forum: EEG data analysis
- Topic: Seeking help with CMS and DRL location with EEGLAB
- Replies: 4
- Views: 29050
Re: Seeking help with CMS and DRL location with EEGLAB
Reference the data. After referencing, the locations of CMS and DRL are not relevant anymore.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Thu Apr 18, 2024 10:29 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Making a config file
- Replies: 1
- Views: 16102
- Sat Mar 09, 2024 5:48 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Event Related Average module
- Replies: 4
- Views: 35779
Re: Event Related Average module
That depends on whether the evoked responses also alternate in polarity. With ABR for example, that's not the case (so the noise trace still shows the noise).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Mar 08, 2024 7:23 pm
- Forum: BioSemi Hardware
- Topic: Active Two Questions (CM. ABR, etc.)
- Replies: 4
- Views: 27692
Re: Active Two Questions (CM. ABR, etc.)
The trigger port needs TTL level signals. In order to connect an audio line signal, you will need to convert to TTL with a comparator and flip-flop. The accuracy of trigger timing is one sample length, see https://www.biosemi.com/faq/review_trigger_timing.htm
Best regards, Coen
Best regards, Coen