Search found 3 matches

by Sajk
Fri Jan 20, 2012 6:26 am
Forum: BioSemi Hardware
Topic: Salt solution problem
Replies: 5
Views: 21127

Thanks for the prompt reply Coen,

Wanted to let you know that we re-ran the tests on 6 electrode sets (3 pairs) and found that the offsets were stable and variation within an electrode set was less than 10mV (no single electrode offset was greater than 10mV).

We found that the "popping" during ...
by Sajk
Wed Jan 18, 2012 11:55 pm
Forum: BioSemi Hardware
Topic: Salt solution problem
Replies: 5
Views: 21127

Thanks Coen that's very reassuring, perhaps we haven't ruined our
electrodes!

We have just started to do the bucket tests as part of regular maintenance
on a number of electrode sets, both individual flat type (EXG) and sets of
32 pin type (with/without CMS/DRL).

We noticed that initially the ...
by Sajk
Fri Jan 13, 2012 6:24 am
Forum: BioSemi Hardware
Topic: Salt solution problem
Replies: 5
Views: 21127

Salt solution problem

We inadvertantly did the bucket test for testing electrodes with iodized salt. Once we realised what we did, we immediately rinsed the electrodes with water and used regular salt but we can't seem to get the impedences down. Has anyone else encountered this, and is there any way to salvage the ...