				Analog to Digital Units
				Posted: Tue Jun 01, 2004 6:53 pm
				by Guest
				What is the Analog to Digital units range that BioSemi uses?  Is this listed anywhere in the documentation or webpage?
Thanks, Jake
			 
			
					
				
				Posted: Tue Jun 01, 2004 9:09 pm
				by Coen
				We use 24-bit sigma-delta analog-to digital-converters (each channels has its own converter), with the preamp gain set for a Least Significant Bit value of 1/32th uV (31.25 nV). This means an input range from +262 mV to -262 mV (+/- 2^23*31.25 nV). More on this topic is found on 
http://www.biosemi.com/faq/smallest_bitstep.htm, 
http://www.biosemi.com/faq/24bitsystem.htm, and 
http://www.biosemi.com/faq/adjust_gain.htm, and 
http://www.biosemi.com/faq/1ADC_per_channel.htm
Best regards, Coen (BioSemi)