Page 1 of 1

what is going on with the raw trigger data?

Posted: Thu Feb 10, 2022 2:41 am
by dmedine
Why does the raw trigger data look like this:
raw_trigger_data.png
raw_trigger_data.png (8.04 KiB) Viewed 45312 times
Close up showing expected shape of trigger with unexpected trigger values (and one of the intermittent huge negative spikes):
closeup.png
closeup.png (13.7 KiB) Viewed 45312 times
I would expect an integer value corresponding to the actual active pins on the parallel port connection. So two questions, what is the meaning of the huge negative floating point numbers and why the huge negative spikes every 5 seconds or so?

Re: what is going on with the raw trigger data?

Posted: Fri Feb 11, 2022 1:14 pm
by Coen
The samples are 24-bit words. The 16 least significant bits are the 16 trigger lines. The 8 most significant bits are used for system status, see https://www.biosemi.com/faq/trigger_signals.htm .

Best regards, Coen (BioSemi)

Re: what is going on with the raw trigger data?

Posted: Sun Feb 13, 2022 10:03 pm
by dmedine
That makes sense. Thanks for pointing me to the FAQ. Somehow I couldn't find the documentation for this.