				sampling rate and bioSemi nyquist
				Posted: Wed Dec 27, 2006 4:16 pm
				by brroach
				BioSemi Users,
What is the maximum resolvable frequency given a sampling rate of X on the BioSemi system with ActiveView2 software?  Is it X/2Hz?  I believe it is X/4Hz on neuroscan systems, and I was curious about how bioSemi compares.
If we are sampling at 1024Hz, can we resolve up to 512Hz?
thanks in advance!
Brian
			 
			
					
				
				Posted: Wed Dec 27, 2006 8:39 pm
				by Coen
				In the ActiveTwo, the-3dB point is at approx 1/5th of the sample rate, also see 
http://www.biosemi.com/faq/adjust_filter.htm. For a selected sample-rate, the ActiView software displays the cutoff frequency in the right sidebar. Competing systems with a 24-bit ADC show exactly the same low-pass response.
Best regards, Coen (BioSemi)
 
			
					
				
				Posted: Tue Jan 23, 2007 12:10 am
				by brroach
				Thanks for your response.  What about the lowest resolvable frequency - does the system place any restriction on that (something like a high pass filter) or is it only based on the amount of data recorded?
thanks again,
Brian
			 
			
					
				
				Posted: Tue Jan 23, 2007 1:45 pm
				by Coen
				The ActiveTwo system uses DC coupled amplifiers, no high-pass filters are applied. The high dynamic range of the used 24-bit ADCs allowed elimination of all coupling capacitors from the design. So, the above statement is correct: the lowest resolvable frequency is only limited by the duration of the experiment.