Hi,
I'm new to BioSemi system and have a question about triggering. Please forgive me if I'm asking about something obvious.
We have no problem with inputting stimulus trigger signals, but we want to use response trigger signals. Can ActiVew accept them and discriminate between stimulus and response signals?

Thank you for any help!
Asuka

Posted: Thu Jul 23, 2009 4:40 pm
by Coen
On the trigger port, 16 input lines are available (see http://www.biosemi.com/faq/trigger_signals.htm). Many researchers use triggers 1-8 for stimuli, and triggers 9-16 for responses. ActiView supports this use by offering trigger display modes (decimal, binary and custom) that show the triggers in two separate groups (each an 8-bit result).

Best regards, Coen (BioSemi)

Thanks!

Posted: Mon Aug 10, 2009 9:43 am
by Asuka
Hi Coen,

Thank you so much!

Asuka