When I import the BDF files into EEGLAB it gives me the wrong number of events. I should have 21 events before I epoch the data, and 20 events after I epoch. The problem lies in that I get the wrong number of events, and they are always different.

The strange thing is that when I put the electrodes in water and run the whole experiment, I get the correct number of events. This only happens when the electrodes aren't recording off an actual human.

Is this a problem with BIOSEMI or EEGLAB? Are there any steps I can take to correct this issue?

Thank you.
Coen
Site Admin
Posts: 1161
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

It must be an EEGLAB problem. In the BioSemi hard- and software, the triggers and EEG data can not influence each other. You can check the files with the simple BDF reader (http://www.biosemi.com/download/BDFReader.zip).