ABR / ERP simultaneous recording

Post Reply
Emz
Posts: 1
Joined: Thu May 12, 2011 8:45 pm
Location: Montreal

ABR / ERP simultaneous recording

Post by Emz »

We are attempting to collect both ABRs and ERPs simultaneously, by recording raw data then offline filtering and post-processing.

We started using the ABR mode and got fairly good results, although our ERPs appeared attenuated and we suppose this was due to the 100Hz high pass filter.

Some colleagues helped us to set up a standard mode recording using the external ports for our 1 data electrode and 2 reference electrodes. We used speedmode7 (though to be honest I'm not entirely sure what the implications of that are). We are still fiddling with our data, but it looks rather noisy.

Is it possible to collect good data for both ABRs and ERPs simultaneously, and if so, what parameters/set up would you recommend?

Emz

Coen
Site Admin
Posts: 1148
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

We did not foresee that customers would want to perform simultaneous ABR and ERP. Therefore, the current firmware and acquisition software does not support simultaneous ABR and ERP.

The possibilities with your current system are:

- Speedmode 9: record ABR only from the special ABR electrodes with the special ABR acquisition software (ActiABR-101)
- Speedmodes 4-7: record with standard active electrodes according to http://www.biosemi.com/faq/adjust_samplerate.htm with standard ActiView.

In speedmode 7, you can acquire channels 1-32 and the 8 EX channels and the sensor channels on 16 kHz. In speedmode 7, the special ABR channels are disabled. It is possible to use 2 standard EX electrodes to record ABR in this speedmode, see http://www.biosemi.com/abr.htm. However, with stadnard EX electrodes the (high frequency) signal-to-noise ratio may be worse than with the special ABR electrodes (because the standard EX electrodes and channels are optimized for lower EEG frequencies, whereas the ABR electrodes and channels are optimized for high frequencies).

Recording the special ABR electrodes together with a number of standard electrodes is technically possible, but would require custom made firmware and acquisition software. Please contact BioSemi directly to discuss the requirements and possibilities.

Best regards, Coen (BioSemi)

tom_campbell75
Posts: 18
Joined: Fri Dec 04, 2009 12:34 am
Location: Center for Visual and Cognitive Neuroscience, North Dakota State University

Recording ABRs, cABRs and MLRs in speed mode 7

Post by tom_campbell75 »

Hi,

I hope that this article is of interest:

http://sccn.ucsd.edu/eeglab/refs/Campbell11.pdf

Best regards,

Tom.

Alexandre
Posts: 13
Joined: Tue Jul 07, 2009 3:22 pm
Location: Montreal

Thank you

Post by Alexandre »

Thank you for this fresh paper. So if I understand well you actually used mode 7 with conventional Biosemi electrodes ? Did you carry out your analyses with EEGLAB ? What about introducing a random jitter between two stimuli to avoid responses from previous trials ?

Best

Alexandre

tom_campbell75
Posts: 18
Joined: Fri Dec 04, 2009 12:34 am
Location: Center for Visual and Cognitive Neuroscience, North Dakota State University

Post by tom_campbell75 »

Hi Alexandre, Yes Mode 7 with the conventional pin electrodes. Sure, I analysed it in MATLAB. Jittering and Adjarring is in the pipeline. T.

sarahah
Posts: 1
Joined: Thu Oct 12, 2017 11:53 am
Contact:

Re: ABR / ERP simultaneous recording

Post by sarahah »

Is there a global value of CFM & ERP? I have found methods to override global erp & cfm values in constraints, but I'd rather like to set the global ones. The problem is I can't find where to set them.