We would like to measure the delay between the event triggers and the stimulus that are show on a projector screen. To accomplish this, we are thinking of using a phototransistor to detect the light intensity that changes on the projector. Can you tell us if we can connect a phototransistor (with some modification) to one of the ADC-06 inputs, and collect the phototransistor output using ActView700?
Can you also provide us the datasheet of ADC-06? Thank you
Regards
Delay Measurement
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
The electrode inputs cannot be used to connect a photo diode or transistor, you will have to order the installation of an Ergo input on one of the Aux connectors on the front-panel of the AD-box (EUR 400). The Ergo input can be used for our standard photo sensor (EUR 150, see http://www.biosemi.com/accessoires.htm), or you can choose to make your own light sensor.
A data sheet for the ActiveTwo system is found on http://www.biosemi.com/activetwo_full_specs.htm
Best regards, Coen (BioSemi)
A data sheet for the ActiveTwo system is found on http://www.biosemi.com/activetwo_full_specs.htm
Best regards, Coen (BioSemi)