use of shocks with BioSemi

Post Reply
jfanning
Posts: 1
Joined: Thu Jan 30, 2014 7:41 pm

use of shocks with BioSemi

Post by jfanning »

Hello, I am interested in running a study in which I administer occassional shocks (range 10-100V, 100Hz, 50-250 ms duration) to subjects' fingertips while recording EEG. I will not be analyzing the EEG data surrounding the shocks. I would like to make sure that this will be safe for both the subject and the equipment, which is the 128-channel setup. The shocks will be administered through two electrodes, one on the first finger and one on the second finger, using Biopac equipment which will have its own ground. According to Biopac their stimulation equipment is safe for use with EEG; however, I would like to verify that this is true even with the active electrodes.

Thank you,

Jennifer
Coen
Site Admin
Posts: 1165
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: use of shocks with BioSemi

Post by Coen »

The above procedure is safe for the ActiveTwo system.

Best regards,Coen (Biosemi)
Post Reply