Hi, everyone,
I have a question on high/low pass settting with online EEG data collecting. specifically speaking, On the monopolar display, what the range of Hz should I set for high/low pass? there is a default low pass (100 Hz) and high pass (0.16Hz). Should tick it or not?
On one participant, I tried disenable the filtering (untick these two boxes), the EEG data is very bad.
What should I do here?
Best,
Jim
bandpass high/low pass
-
Coen
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: bandpass high/low pass
The filters (and reference selection) only affect the displayed signals (not the data to file, that's always DC - 1/5th of the sample rate, and unreferenced).
Set the filters to limit the display to the bandwidth of interest (suppress low frequency drift and high frequency noise). Choose a proper reference for sufficient rejection of Common Mode on the displayed signals.
Best regards, Coen (BioSemi)
Set the filters to limit the display to the bandwidth of interest (suppress low frequency drift and high frequency noise). Choose a proper reference for sufficient rejection of Common Mode on the displayed signals.