external electrode with high impedance

Post Reply
mpwa
Posts: 2
Joined: Mon Mar 16, 2015 10:56 am

external electrode with high impedance

Post by mpwa »

Hello,

One of our external electrodes has a constant high impedance. it is not the AD box, because if we use another external electrode, its impedance is fine. Is there anyway to fix this? the CM stays in range, that is the AD box has a solid blue light, so none of the other electrodes are affected.

Thanks
Martin
Coen
Site Admin
Posts: 1161
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: external electrode with high impedance

Post by Coen »

It's a high offset, not impedance.

If within +50 mV and - 50 mV and stable, then all is OK. If not, replace the electrode.

Best regards, Coen (BioSemi)
Post Reply