Analog to Digital Units

Post Reply
Guest

Analog to Digital Units

Post by Guest »

What is the Analog to Digital units range that BioSemi uses? Is this listed anywhere in the documentation or webpage?
Thanks, Jake
Coen
Site Admin
Posts: 1164
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

We use 24-bit sigma-delta analog-to digital-converters (each channels has its own converter), with the preamp gain set for a Least Significant Bit value of 1/32th uV (31.25 nV). This means an input range from +262 mV to -262 mV (+/- 2^23*31.25 nV). More on this topic is found on http://www.biosemi.com/faq/smallest_bitstep.htm, http://www.biosemi.com/faq/24bitsystem.htm, and http://www.biosemi.com/faq/adjust_gain.htm, and http://www.biosemi.com/faq/1ADC_per_channel.htm

Best regards, Coen (BioSemi)
Post Reply