Lowpass filter
cbrnr
Lowpass filter
Post by cbrnr »
Hello!
I'm a bit confused about the lowpass filter settings used during recording. I have an ActiveTwo amp, and I usually set the speed dial mode to 4 (sampling rate 2048Hz). In ActiView, I select a decimation factor of 1/4 to get signals with a sampling rate of 512Hz.
According to your website, the cutoff frequency of the hardware lowpass filter is 1/5 of the sampling rate, so 2048Hz / 5 = 409.6Hz for my setting. After decimation, this should be 409.6Hz / 4 = 102.4Hz. However, the BDF file headers contain a field with "LP: 104Hz" (which is also what ActiView shows for the bandwidth).
What is the correct cutoff frequency?
Clemens
Re: Lowpass filter
1/5th is an approximation, 104 Hz is the correct -3dB frequency, for the response of the CIC filter see https://www.biosemi.com/faq/adjust_filter.htm
cbrnr
Re: Lowpass filter
Post by cbrnr »
Thanks Coen, it's good to know that I can trust the BDF header!
