BioSemi Users,
What is the maximum resolvable frequency given a sampling rate of X on the BioSemi system with ActiveView2 software?  Is it X/2Hz?  I believe it is X/4Hz on neuroscan systems, and I was curious about how bioSemi compares.
If we are sampling at 1024Hz, can we resolve up to 512Hz?
thanks in advance!
Brian
			
			
									
						
										
						sampling rate and bioSemi nyquist
- 
				Coen
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
In the ActiveTwo, the-3dB point is at approx 1/5th of the sample rate, also see http://www.biosemi.com/faq/adjust_filter.htm. For a selected sample-rate, the ActiView software displays the cutoff frequency in the right sidebar. Competing systems with a 24-bit ADC show exactly the same low-pass response.
Best regards, Coen (BioSemi)
			
			
									
						
										
						Best regards, Coen (BioSemi)
- 
				Coen
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
The ActiveTwo system uses DC coupled amplifiers, no high-pass filters are applied. The high dynamic range of the used 24-bit ADCs allowed elimination of all coupling capacitors from the design. So, the above statement is correct: the lowest resolvable frequency is only limited by the duration of the experiment.
Best regards, Coen (BioSemi)
			
			
									
						
										