Ergo port, treshold voltage for trigger
Ergo port, treshold voltage for trigger
I couldn't find anything about what voltage the Ergoport triggers when configured as trigger. Can I also adjust this value?
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Ergo port, treshold voltage for trigger
The one-shot trigger characteristics can be adjusted in the CFG (use ActiTools --> Configurator). Further explanation is found at https://www.youtube.com/watch?v=_OMoXttIguc&t=560s
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
Re: Ergo port, treshold voltage for trigger
Thanks for the information. I have a biosemi with two ergoports and connected two sensors. Sometimes I get a lot of noise (200mv pp) at ergoport 1 but after waiting a few minutes, the noise disapears, while ergoport 2 is working perfectly fine. I used +3v and -2v to power the sensor, and 0v is not connected. Switching the two sensors gives me the same result, aka the problem is not caused by the sensors. I measure between +10 and +200mv.
Oh, and I checked the plug thoroughly and resoldered everything to make sure it's well connected.
Oh, and I checked the plug thoroughly and resoldered everything to make sure it's well connected.
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Ergo port, treshold voltage for trigger
Sounds like a missing/floating 0V/Ground. Sensor output should probably be referenced to 0V. Connect the 0V to the sensor Ground/Common. Supply datasheet of sensor.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
Re: Ergo port, treshold voltage for trigger
sensor type:https://www.omega.com/en-us/pressure-me ... V5(150psi)
Sensor messures between .5 volts and 4.5, measuring 0-2 bar. Supply of the sensor is connected between +3v and -2v. Input of Biosemi is connected via a 2k7 resistor to the sensor and via a 5k6 resistor to +3v. There's two ergo ports in the enclosure. Port 2 works fine, port 1 does not.
Sensor messures between .5 volts and 4.5, measuring 0-2 bar. Supply of the sensor is connected between +3v and -2v. Input of Biosemi is connected via a 2k7 resistor to the sensor and via a 5k6 resistor to +3v. There's two ergo ports in the enclosure. Port 2 works fine, port 1 does not.
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Ergo port, treshold voltage for trigger
Did you connect the inverting inputs of the ERGOs (pins 5 and 7) to 0V/Common (pin 2) ?
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
Re: Ergo port, treshold voltage for trigger
No, I didn't. Should I?
Is there any way to view the y axis in absolute voltage in active view?
Is there any way to view the y axis in absolute voltage in active view?
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Ergo port, treshold voltage for trigger
Yes you should, unconnected inverting inputs could explain the above problems. Inverting inputs to 0V is OK, but in your case is even better to connect the inverting inputs via a 5k6 resistor to +3V, AND via a 3k7 resistor to -2V (better rejection of supply voltage ripple because both inverting and non-inverting inputs reference to the supply rails).
Absolute values can be seen in ActiRead ---> Value (http://www.biosemi.com/download/ToolsFo ... 02-Win.zip), but not in ActiView.
Best regards, Coen (BioSemi)
Absolute values can be seen in ActiRead ---> Value (http://www.biosemi.com/download/ToolsFo ... 02-Win.zip), but not in ActiView.
Best regards, Coen (BioSemi)
Re: Ergo port, treshold voltage for trigger
Thanks, that worked!
Re: Ergo port, treshold voltage for trigger
Is there any method to adjust the trigger threshold during runtime?
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Ergo port, treshold voltage for trigger
ActiView is open-source. You can change the LabVIEW code to not read the trigger level from the CFG, but from an added adjuster.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)