Speed-mode setting on BioSemi AD box

Post Reply
hkeyes
Posts: 5
Joined: Fri Nov 11, 2005 3:16 pm

Speed-mode setting on BioSemi AD box

Post by hkeyes »

Hello,

We have just purchased a 64-channel BioSemi system. I am unsure as to which speed-mode setting I should be using. I will be recording from 64 electrodes plus 8 flat-type active electrodes (for EOG). Can anyone tell me what the most suitable speed-mode setting would be for this set-up?

From the table in the manual, it appears that speed-mode setting 4 is appropriate- 2048 (2kHz), 232 + 8. In the sensors box, it says "yes". Does "sensors" in this case refer to the 8 EOG electrodes?

Thanks,

Helen Keyes

School of Psychology
UCD Dublin
Coen
Site Admin
Posts: 1154
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

You can use speedmodes 4, 5 or 6. Speedmode 6 will give you a maximum sample-rate of 8 kHz, and a minimum sample rate of 128 Hz (by using the "decimation" option in software).

"Sensors" refers to the extra optional sensors connected to the Aux inputs (GSR, Plethysmograph, Respiration, ect, see http://www.biosemi.com/accessoires.htm).

Best regards, Coen (BioSemi)
hkeyes
Posts: 5
Joined: Fri Nov 11, 2005 3:16 pm

Post by hkeyes »

Thanks very much Coen, that's really helpful.

Helen
Coen
Site Admin
Posts: 1154
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

Just one extra remark: keep in mind that the analog bandwidth is approx. 1/5th of the sample-rate (-3 dB point is indicated in the right sidebar). So, ensure that you select the sample-rate high enough to retain sufficient high frequency information in the data. For most EEG work, it is recommended to not go lower than 512 Hz (-3 dB cutoff of the anti-aliasing filter is at 104 Hz). By the way, you can always downsample the data off-line with the "decimator" tool, available from http://www.biosemi.com/download/Decimator.zip.

Best regards, Coen (BioSemi)
Last edited by Coen on Tue Sep 02, 2008 1:02 pm, edited 1 time in total.
hkeyes
Posts: 5
Joined: Fri Nov 11, 2005 3:16 pm

Post by hkeyes »

Thanks Coen,

Helen
Post Reply