Hi there,

I am running a frequency tagging study. The paradigm involves flashing visual stimuli on the screen at 2.5 & 3.16Hz to induce periodic neural responses. I've attached the grand averaged frequency spectrum showing the responses to our stimulation (fundamentals & harmonics highlighted) as well as peaks completely unrelated to our stimulation (unhighlighted). The source appears to be at 1Hz with the rest being its harmonics. I am not sure where this could be coming from as it doesn't align with our stimulation.

I have also included the topography of this 1Hz response which does not look visually based. It looks like it is right above CMS/DRL. Is it possible that something in the hardware is causing this?

Sidenote: there is an additional trigger in some of my files '4', that I did not code into the study. Unsure if this could be related.

topo.JPG
topo.JPG (39.29 KiB) Viewed 64490 times
freqspect.JPG
freqspect.JPG (47.46 KiB) Viewed 64490 times
Best,
Anna
Coen
Site Admin
Posts: 1165
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: 1Hz Unexpected Frequency in Signal

Post by Coen »

"Is it possible that something in the hardware is causing this?"

No, the BioSemi system does not generate any signal of exactly 1 Hz.

Best regards, Coen (BioSemi)