Trace randomly fell completely flat
-
- Posts: 2
- Joined: Sat Jan 29, 2022 2:50 pm
Trace randomly fell completely flat
I was wondering if anyone knows of why/has had a similar situation. I was running an experiment and everything was normal until the trace from all the electrodes randomly went completely flat. It showed no change at any point after that. The electrode offsets remained the same and nothing was changed about the hardware. Do you think this is a bug with the software?
-
- Site Admin
- Posts: 1148
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Trace randomly fell completely flat
Did you press the keyboard shortcut PgUp or PgDn with an earlier version of ActiView than 9.00 ? If yes, then the display switched to a different (empty) group of channels. BDF file is not affected.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
-
- Posts: 2
- Joined: Sat Jan 29, 2022 2:50 pm
Re: Trace randomly fell completely flat
Thanks for your reply Coen.
I am using the newest version of Actiview. My participants could have pressed a button like up or down. Are there any other keyboard prompts I should know about/ where can we find info on this?
I am using the newest version of Actiview. My participants could have pressed a button like up or down. Are there any other keyboard prompts I should know about/ where can we find info on this?
-
- Site Admin
- Posts: 1148
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Trace randomly fell completely flat
Version 9.00 (latest) has the following keyboard shortcuts:
- ENTER = Start
- HOME = Reset baselines (display only)
- F1-F8 = Triggers 9-16
All other keyboard shortcuts (Up/Down, Left/Right, PgUp/PgDn) that were available in older versions of ActiView have been deleted in version 9.00.
Best regards, Coen (BioSemi)
- ENTER = Start
- HOME = Reset baselines (display only)
- F1-F8 = Triggers 9-16
All other keyboard shortcuts (Up/Down, Left/Right, PgUp/PgDn) that were available in older versions of ActiView have been deleted in version 9.00.
Best regards, Coen (BioSemi)