I attempted to perform real-time processing of EEG signals using LabStreamingLayer (LSL), but encountered an issue where the Active3 system could not link to LSL, resulting in the error: "Could not link to the amplifier: Ring buffer handshake with BioSemi driver failed unexpectedly."
Is there a way to resolve this error? If not, are there alternative methods for real-time processing of EEG signals from the Active3 system using MATLAB programming?
Real-time processing of EEG signals from Active3
-
- Posts: 2
- Joined: Fri Nov 15, 2024 10:06 am
-
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Real-time processing of EEG signals from Active3
ActiView for ActiveThree can stream the data from the ringbuffer via a TCP/IP connection. You can consider to develop a MatLab client to read this real-time data stream.
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
-
- Posts: 2
- Joined: Fri Nov 15, 2024 10:06 am
-
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Real-time processing of EEG signals from Active3
https://www.linkedin.com/posts/neurospe ... 3t821cL4Pw
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon.
Best regards, Coen (BioSemi)
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon.
Best regards, Coen (BioSemi)