Hi Coen and the rest of the BioSemi team.
I have been making some measurements using a plethysmograph (and later emg and gsr). I have been picking up the data with TCP, converted to 24-bit numbers and so on and everything looks fine except that sometimes the values are being subtracted by what looks like a certain number.
When I plot it it looks like this:
[attachment=2]figure_1.png[/attachment]
Which is very weird looking. If you notice the top it looks signal the signal that I'm interested in.
[attachment=1]figure_2.png[/attachment]
Which is very similar to what I saw in the actiview acqusition program.
If I remove all values below zero I get this:
[attachment=0]figure_3.png[/attachment]
Which is ultimately what I want, except that almost half of the values are removed.
What causes this? And how can I avoid it?
Weird TCP readings.
-
dunguyen
Weird TCP readings.
- Attachments
-
- figure_3.png (76.55 KiB) Viewed 26790 times
-
- figure_2.png (47.86 KiB) Viewed 26790 times
-
- figure_1.png (131.71 KiB) Viewed 26790 times
-
Coen
- Site Admin
- Posts: 1165
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Weird TCP readings.
Treating two's complement numbers as unsigned ?
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
-
dunguyen
Re: Weird TCP readings.
Thanks that worked!
For anybody else wondering I used this solution: http://stackoverflow.com/a/3783679
The readings are correct now:
[attachment=0]fixed!.png[/attachment]
For anybody else wondering I used this solution: http://stackoverflow.com/a/3783679
The readings are correct now:
[attachment=0]fixed!.png[/attachment]
- Attachments
-
- fixed!.png (77.84 KiB) Viewed 26763 times