Decimation and E-prime triggers

Post Reply
Sam
Posts: 1
Joined: Wed Apr 23, 2014 4:43 am

Decimation and E-prime triggers

Post by Sam »

Hi,

I have a question about decimation affecting the triggers being send through e-prime. When I don't use decimation (i.e decimation: 1) I don't have any problems, but when I want to down sample the triggers change. The triggers changed both when I used decimation with speed mode 5 (when sample rate is 4096) and with speed mode 4 (when sample rate is 2048). It doesnt seem to be the sampling rate itself, since both 4096 Hz and 2048 Hz work fine if I don't use decimation. I added a .jpg to clarify the change.


Thanks in advance,
Sam.
Attachments
triggerprob.jpg
triggerprob.jpg (230.89 KiB) Viewed 26505 times
Coen
Site Admin
Posts: 1165
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Decimation and E-prime triggers

Post by Coen »

During decimation, an OR function is applied to the individual trigger bits within a group of samples. So, do not use decimation when the time between triggers is less than the sample time of the decimated sample rate.

Best regards, Coen (BioSemi)