Confirmed. The channels are always streamed in the same order (regardless of the channel labels and regardless of connected electrode sets), in your case channel 1 to channel 72. The CFG just determines which label is assigned to which channel.
Best regards, Coen (BioSemi)
Search found 1161 matches
- Thu Dec 04, 2025 10:02 pm
- Forum: BioSemi all other topics
- Topic: Number of channels streamead by Biosemi to Lab Streaming Layer
- Replies: 5
- Views: 39075
- Tue Nov 25, 2025 7:43 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView103 sending multiple LSL streams
- Replies: 6
- Views: 79537
Re: ActiView103 sending multiple LSL streams
There was indeed a bug in ActiView that could lead to the generation of multiple LSL streams on some computers, sometimes. The bug is repaired in version 10.4 (https://www.biosemi.com/download/LatestActiView/ActiView104/ActiView104.zip). When no LSL connection is used, this bug could still lead to ...
- Tue Nov 25, 2025 7:31 pm
- Forum: BioSemi all other topics
- Topic: Number of channels streamead by Biosemi to Lab Streaming Layer
- Replies: 5
- Views: 39075
Re: Number of channels streamead by Biosemi to Lab Streaming Layer
That's the way it is currently programmed: all the channels (electrodes+connected sensors+triggers) are always streamed to LSL. In your case, you have 64+8 channels installed in the AD-box and no sensors connected, so 73 channels are streamed. The idea is that you can select a subset in the LSL ...
- Wed Nov 12, 2025 7:08 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView103 sending multiple LSL streams
- Replies: 6
- Views: 79537
Re: ActiView103 sending multiple LSL streams
I have made some changes to the LabVIEW code that should prevent the multiple streams. Please email to info@biosemi.com and I will sent you a download link for a preliminary version of the upcoming ActiView 104.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Nov 12, 2025 3:31 pm
- Forum: EEG data analysis
- Topic: Gain and offset for physical values
- Replies: 7
- Views: 128289
Re: Gain and offset for physical values
BDF files by A2 and A3 are basically the same, you can ignore the offset parameter. Scaling for the EEG channels is the same (LSB = 31.25 nV = 1/32 uV).
A difference between A2 and A3 BDF files is the scaling of some of the auxilary sensors, as defined in the BDF header.
In addition, the analog ...
A difference between A2 and A3 BDF files is the scaling of some of the auxilary sensors, as defined in the BDF header.
In addition, the analog ...
- Mon Nov 10, 2025 5:47 pm
- Forum: BioSemi all other topics
- Topic: Conductive floor
- Replies: 1
- Views: 14809
Re: Conductive floor
The ESD floor has a high resistance (MegaOhm to GigaOhm) and would not make any difference to the interference rejection of the BioSemi system.
The inputs of the BioSemi system are protected against static discharge (standard EN 61326). An ESD floor is not required to prevent damage. Just remain ...
The inputs of the BioSemi system are protected against static discharge (standard EN 61326). An ESD floor is not required to prevent damage. Just remain ...
- Fri Oct 31, 2025 1:11 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView103 sending multiple LSL streams
- Replies: 6
- Views: 79537
Re: ActiView103 sending multiple LSL streams
Does it work with the BrainVision LSL viewer ( https://labstreaminglayer.readthedocs.i ... ewers.html ) ?
Which errors do you see in ActiView (popup messages and/or error counters in the lower right corner) ?
Best regards, Coen (BioSemi)
Which errors do you see in ActiView (popup messages and/or error counters in the lower right corner) ?
Best regards, Coen (BioSemi)
- Mon Jun 16, 2025 9:28 pm
- Forum: BioSemi all other topics
- Topic: Real-time processing of EEG signals from Active3
- Replies: 3
- Views: 116383
Re: Real-time processing of EEG signals from Active3
https://www.linkedin.com/posts/neurospec_news-from-the-house-of-biosemi-with-the-activity-7340398038353416193-BpTW?utm_source=share&utm_medium=member_desktop&rcm=ACoAAAKVmwQBf6tXt0CAtOT6Bv7Kx3t821cL4Pw
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon ...
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon ...
- Mon Jun 09, 2025 1:52 pm
- Forum: BioSemi Hardware
- Topic: Simultaneous EEG + EMG
- Replies: 1
- Views: 23006
Re: Simultaneous EEG + EMG
You should only connect one CMS/DRL to the AD-box (otherwise, the blue light keeps flashing). Connect either the A set with CMS/DRL, or the front-panel CMS/DRL. So, you can use the EX electrodes with the front-panel CMS/DRL provided the A connector remains unplugged.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Apr 02, 2025 1:39 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Electrode labels when checking offsets in ActiView
- Replies: 1
- Views: 48027
Re: Electrode labels when checking offsets in ActiView
Use the CFG file for 64 channels with 1020 labeling, best regards, Coen (BioSemi)
- Tue Feb 11, 2025 10:21 pm
- Forum: BioSemi Hardware
- Topic: Updating firmware for high speed
- Replies: 2
- Views: 74861
Re: Updating firmware for high speed
Contact info@biosemi.com
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
- Tue Feb 04, 2025 3:52 pm
- Forum: BioSemi Hardware
- Topic: Need Advise!!
- Replies: 1
- Views: 26687
Re: Need Advise!!
Since the skin is not abraded, disinfection is generally not needed. When disinfection is preferred (as was the case during the Covid-19 outbreak), then follow the instructions as shown in https://www.youtube.com/watch?v=-CZx3_tbrYM . Do not use hydrogen peroxide since it may damage the headcap ...
- Tue Feb 04, 2025 3:35 pm
- Forum: BioSemi Hardware
- Topic: Compatibility with Transcutaneous vagus nerve stimulation (tVNS)
- Replies: 1
- Views: 27093
- Tue Dec 17, 2024 7:50 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Access Offsets from.bdf files
- Replies: 1
- Views: 61337
- Fri Nov 15, 2024 7:45 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 115585
Re: Error 5000
Contact your local BioSemi representative
- Fri Nov 15, 2024 6:36 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 115585
Re: Error 5000
ADC6 is an Mk1 AD-Box and does not work with the red receiver. Your white receiver does work with the USB Trigger Interface (the trigger ports of white and red receivers are identical).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Nov 15, 2024 1:44 pm
- Forum: BioSemi all other topics
- Topic: Real-time processing of EEG signals from Active3
- Replies: 3
- Views: 116383
Re: Real-time processing of EEG signals from Active3
ActiView for ActiveThree can stream the data from the ringbuffer via a TCP/IP connection. You can consider to develop a MatLab client to read this real-time data stream.
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
- Tue Nov 12, 2024 5:59 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 48503
Re: ActiView10 not working with USB2 receiver with Firmware4.0
It is normal operation that only trigger lines 15 (start/stop) and 16 (1 sec pulses) are transmitted from Master to Slave. The stimulation triggers should indeed be connected to the Master. The trigger lines 1-14 of the Slave are ignored when the Master and Slave files are merged into a single ...
- Mon Nov 11, 2024 3:01 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 48503
Re: ActiView10 not working with USB2 receiver with Firmware4.0
ActiveTwo (Mk2 and Mk2.5) only works with ActiView 9.xx
ActiveThree only works with ActiView 10.x
Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling:
Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate
Slave: ActiveTwo with ActiView 9.03 on 2 kHz ...
ActiveThree only works with ActiView 10.x
Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling:
Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate
Slave: ActiveTwo with ActiView 9.03 on 2 kHz ...
- Wed Nov 06, 2024 6:30 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Errors 7000 and 7001
- Replies: 2
- Views: 32355
Re: Errors 7000 and 7001
The above CFG is for ActiveTwo.
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
- Thu Oct 31, 2024 3:41 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 40000
Re: hardware filters
Which "analog box" exactly are you talking about ?
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
- Thu Oct 31, 2024 3:25 pm
- Forum: BioSemi all other topics
- Topic: BIOSEMI Trigger and EEGLAB reading events
- Replies: 1
- Views: 34562
Re: BIOSEMI Trigger and EEGLAB reading events
ActiveTwo Mk1 and Mk2 (as well as the new ActiveThree) systems use exactly the same BDF file format (https://www.biosemi.com/faq/file_format.htm). In particular, storing of the 16 trigger lines in the Status channel is the same for all systems. The only difference between Mk1 and Mk2 files is the ...
- Thu Oct 24, 2024 6:38 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Older Versions Repository? Looking for ActiView Version 705
- Replies: 1
- Views: 27854
Re: Older Versions Repository? Looking for ActiView Version 705
Use the current 9.02 version, it works with every ActiveTwo system (even as old as from 2002). You might have to upgrade the firmware of your USB-receiver to use it with Windows 10/11, see https://www.biosemi.com/faq/install_USB.htm.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Oct 02, 2024 1:52 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 40000
Re: hardware filters
The BioSemi system has DC amplifiers, there is no high-pass filter. Check the signal in ActiView with the (display) filters switched off. You have either a high-pass in your file reader, or you have capacitively coupled the trigger line to the Analog Input Box.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Sep 25, 2024 1:32 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 55795
Re: Hyperscanning with LabStreamingLayer
With the triggered hyperscanning method, the data files from the two (or more) systems are merged into a single hyperscanning file offline. So it is not possible to stream hyperscanning data during the recording.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Sep 24, 2024 3:19 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 55795
Re: Hyperscanning with LabStreamingLayer
According to the LSL website it should work with ActiveTwo (2 and 2.5), but that's all I know.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Mon Sep 23, 2024 2:30 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 55795
Re: Hyperscanning with LabStreamingLayer
You can Hyperscan with the Active3 and Active2.5 system using the method described on https://www.biosemi.com/faq/Hyperscanni ... erface.htm .
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
- Mon Sep 09, 2024 6:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 115585
Re: Error 5000
Send the red receiver to BioSemi, we will replace it with a white one (we still have a few).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 5:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 115585
Re: Error 5000
ADC11 are Mk1 systems (20 years old), they do not work with the new red receiver. You should use the older type white receiver.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 4:15 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 115585
Re: Error 5000
What is the serial number of your AD-box ?
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)