Hello 
I would like to ask what is the recommended procedure and clearing and disinfecting agent for: (1) EEG cap; (2) Cap electrode; (3) EMG flat electrodes; and (4) GSR sensor?
Is it possible to use alcohol?
Thanks
Bori
			
			
									
						
										
						How to disinfect Biosemi hardware?
- 
				Coen
- Site Admin
- Posts: 1154
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: How to disinfect Biosemi hardware?
Disinfection is not necessary because it is a non-invasive system (no skin scrubbing is applied to lower skin impedances).
If you insist on disinfection then we recommend:
- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps
Best regards, Coen (BioSemi)
			
			
									
						
										
						If you insist on disinfection then we recommend:
- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps
Best regards, Coen (BioSemi)
Re: How to disinfect Biosemi hardware?
Thank you very much. Is the percent of the alcohol meters?
Best
Boris
			
			
									
						
										
						Best
Boris