I would like to ask what is the recommended procedure and clearing and disinfecting agent for: (1) EEG cap; (2) Cap electrode; (3) EMG flat electrodes; and (4) GSR sensor?
Is it possible to use alcohol?
Thanks
Bori
How to disinfect Biosemi hardware?
-
Coen
- Site Admin
- Posts: 1161
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: How to disinfect Biosemi hardware?
Disinfection is not necessary because it is a non-invasive system (no skin scrubbing is applied to lower skin impedances).
If you insist on disinfection then we recommend:
- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps
Best regards, Coen (BioSemi)
If you insist on disinfection then we recommend:
- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps
Best regards, Coen (BioSemi)
Re: How to disinfect Biosemi hardware?
Thank you very much. Is the percent of the alcohol meters?